这是摘自EN285-2006的内容,已解释的非常详细,请楼主参考。
7.1.14 An automatic test cycle shall be provided to carry out the air leakage test (see Clause 18).The error in measurement for a pressure difference of any 1,5 kPa (15 mbar) shall not exceed 0,1 kPa (1 mbar) over the pressure range that may occur during the test.
18.2.1 Test pressure instrument as described in 24.3.
If the sterilizer is fitted with an absolute pressure instrument complying with 24.3 this additional gauge is not
required
24.3.2 The resolution shall not exceed 1 kPa over the range 0 kPa to 400 kPa and for instruments used for air
leakage test the resolution shall not exceed 0,1 kPa over the range used [see 24.3.3 b)].
24.3.3 The test measurement system shall:
a) have a scale range which includes 0 kPa to 400 kPa (-1 bar to 3 bar) and an accuracy class over this range of
at least 0,25;
b) if used for the air leakage test have a scale range specified by the manufacturer of the sterilizer and have an
accuracy of not less than 0,1 kPa (1 mbar) over any pressure difference of 1,5 kPa (15 mbar) over the
pressure range that may occur during the test.
|
|