In a clinical use environment, the health care provider should wear protective disposable gloves and cover the Sensor with a hygienic barrier. Before using the Sensor the first time, and before every new patient, the following protocol is recommended: Remove and discard all protective hygienic barriers and / or sheaths from the Sensor prior to removing disposable gloves.
Place the Sensor on a tray covered by a disposable liner, or in a receptacle that can be thoroughly cleaned.
Remove and discard gloves.
Wash hands and put on a new pair of disposable gloves.
Disconnect Sensor from USB Interface.
If the Sensor or cable are visibly soiled (e.g., with blood or saliva), each should be cleaned with a soapy cloth or paper towel, and then dried with a clean lint-free cloth or paper towel.
Thoroughly spray or wipe the Sensor and cable with 75% ethanol. Do not expose the Sensor cable connectors to any amount of liquid. If using a spray detergent, allow it to remain on the Sensor for 5 minutes.
If using a liquid detergent, allow it to remain on the Sensor for 30