8.59. Validation of porous cycles should include a calculation of equilibration time, exposure time, correlation of pressure and temperature and the minimum/maximum temperature range during exposure. Validation of fluid cycles should include temperature, time and/or F0. Critical processing parameters should be subject to defined limits (including appropriate tolerances) and be confirmed as part of the sterilisation validation and routine cycle acceptance criteria.