December 2005Selecting the Swab Sampling Area
This Cleaning Memo addresses the question of the specific surface area to swab in a cleaning validationprotocol. For clarification, this will address the size of swab area, not the location of the swab area. Mostcompanies will swab areas of 25 cm2(4 square inches) to 100 cm2(16 square inches), although I have seenswab areas as large as 400 cm2.
There are no regulatory guidelines for selecting the size of the area to be sampled. Note, however, that therehas been at least one FDA 483 citation for using too small an area for sampling environmental surfaces forpenicillin. It should be noted that this concern was specific to this situation, in which it was probably believedthat swabbing a smaller area was less likely to detect penicillin contamination on environmental surfaces.
What are the advantages and disadvantages of different swab areas for swabbing? For simplicity, we willconsider sampled areas of 25 cm2as a small area and 100 cm2as a larger area. In each point discussed below,it will be considered on an “other things being equal” basis.
One advantage of swabbing a larger area is that the residue limit in the analytical sample will be higher. Thismeans that the limit of detection or limit of quantitation of the analytical limit could be higher. For example, ifI swab 100 cm2and then desorb it into 20 mL of solvent, the residue limit in the analytical sample will be fourtimes larger than the residue limit if only 25 cm2 were sampled and the swab desorbed into 20 mL of solvent.Note that while the limit in the analytical sample is higher with the larger surface area, there is no difference interms of the risk of passing or failing (because the amount of residue potentially swabbed from the surface willbe four times as great with the larger surface area).One advantage of swabbing a smaller area is that I am more likely to get a higher recovery. As a generalprinciple, the larger the surface area the lower the recovery. This may be due to the fact that when I use a wetswab (as is usually the case), I am going to leave a finite amount of liquid on the swabbed surface. The largerthe area swabbed, the more liquid will be left behind. It should be noted that the difference may be small incomparison to the overall recovery. One way to minimize this effect is to use two swabs on the same surfacearea, with the second swab being dry.
Another factor in the use of a larger area is that there may be less variability in the actual area sampled. If I amoff by a length of 0.5 cm in each direction for my swabbing, then the variability of area coverage may be asmuch as ± 10% for a swab area of 100 cm2. If I am off by the same 0.5 cm in each direction for a swab area of25 cm2, then the variability of area coverage may be as much as ± 20%. This effect can be negated if I were touse a template for controlling the sampled area (instead of using the “eyeball” method).
A final consideration is that by selecting the smaller surface area (25 cm2), the more likely it is that allsampling locations will have that area available. With a larger surface area of 100 cm2, the more likely it willbe that there will be certain locations where I can’t sample a full 100 cm2(for example, perhaps only 80 cm2isavailable for swabbing). In addition, the smaller the swabbed area the more likely it will be that I will be ableto sample a location that would be square (for example, 10 cm X 10 cm as opposed to 5 cm X 20 cm). Thisissue is situational, and may or may not be significant. Note also that I am not addressing the issue of how to handle recovery if a different swabbing shape is used. I am just stating that the situation of having to use a different shape (other than square) is more likely to occur the larger the swabbed area.
These are some of the considerations in selecting the size of the swabbed area in a validation protocol.Needless to say, for training and consistency reasons, one should generally select one specific size forswabbing in protocols and in recovery studies.