ISPE第一版C&Q 征求意见时,FDA的comments 如下:
In Chapter 11, item 2 Standard Operation Procedures section fails to describe when the SOPs for the operation, maintenance, and calibration of equipment should be completed. Earlier sections reference SOPs, but do not specifically address the operational, maintenance, and calibration SOPs. These SOPs should be written and finalized before or during the OQ process, so they can be used and referenced in the performance qualification. We think a little bit of clarification on this issue would be helpful. It is also important to have SOPs for cleaning, and we believe these should be developed during the OQ process. The operational qualification protocol could require as part of it’s final acceptance criteria that all of these important SOPs be completed and approved. |
|